KNOW YOUR NUMBERS
By: Hill Cox
A manufacturer using outside, contracted calibration services needs to
pay close attention to the pretty certificates he receives from the laboratory.
The problem is that measurement uncertainty statements on certificates often show the same
values as the Best Measurement Capability on the Scope of Accreditation issued by the
laboratory's accrediting agency.
Simply put, uncertainty numbers on the Scope refer to a measurement in the laboratory on an item
in pristine condition. The laboratory will use the highest reference standards and most accurate
instruments they have for such tests, not necessarily the ones used for everyday work.
If the gage you send for calibration is not mangled or rusty, and the laboratory uses the same
masters and equipment to calibrate it that were used in tests to support their uncertainty
claims, then the uncertainty numbers could be the same as those on the lab-oratory's Scope of
Accreditation. The odds are they won't be the same as the numbers that are listed, because if
the laboratory calibrated your gage the way they did for the test they submitted for
accreditation, you'd be paying much more for their everyday service.
When it comes to on-site calibration, laboratory conditions are rarely duplicated. Thus, there
may be a big difference between the uncertainty on the report given to you as compared to the
value shown on the Scope of Accreditation. Don't blame the calibration laboratory if the
uncertainty numbers for such work end up in the twilight zone.
If you have concerns about the uncertainty numbers reported to you, talk to the
laboratory that did the work. When on-site work is involved and the calibration laboratory is
using the same values as shown on their Scope of Accreditation, ask them for the temperature
range for which the uncertainties are valid. If the response you receive is something such as,
"One-half a degree," or some other value that seems to be only obtainable in a
laboratory setting, and you know the temperature in your facility fluctuates by several degrees,
it's time to take a close look at the calibration laboratory's uncertainty budgets that they
report to you and determine what else may not be quite right.
My comments are based on the assumption that the uncertainty values on laboratory Scopes reflect
reality of one kind or another. Unfortunately, many don't. You knew that was coming, didn't you?
Some organizations accredit laboratories whose Scopes are deserving of awards for creative
writing but bear little resemblance to real-life calibration. Such laboratories are able to
accomplish this because organizations that offer cheap and fast accreditation are able to do so
by focusing on quality systems using submitted paperwork only. Such accreditation agencies
rarely go beyond what is written on paper and verify the laboratory's results independently. The
technical data for the Scope of Accreditation is often ignored, or the assessor has little or no
knowledge or experience and accepts anything that looks reasonable as an uncertainty claim.
What's the big deal? It may be no big deal if you're only interested in paperwork for your files.
But if you need to prepare uncertainty budgets for your everyday measurements-and every
manufacturer should, while others must for the sake of fulfilling standards requirements-those
must include the uncertainties for the measuring equipment. If those uncertainties don't reflect
reality, your measurement uncertainty budgets won't either and everything goes downhill from
there.
Reputable accrediting agencies scrutinize reports issued by laboratories they have accredited.
This ensures that correct practices are being followed in reporting measurement uncertainties.
Unfortunately, some of those who accredit or register calibration laboratories may not be as
diligent. If you want to ensure the numbers you receive reflect the reality you'll have to live
with, carefully review the reports you receive and the Scopes behind them. And, if necessary,
evaluate the accrediting agency endorsing them.
Hill Cox
President
The American Measuring Tool Manufacturers Association (AMTMA)
Used
with permission from Quality Magazine, March
2003 issue. Hill Cox is president of The American
Measuring Tool Manufacturers Association (AMTMA)
and is chair of their technical committee. He is also
president of Frank J. Cox Sales Ltd. (Brampton,
Ontario, Canada).
Comment on this article
|
|
|